I am getting this error when removing instrument response and was wondering if anyone could help with a solution:
instrument response deconvolution…
EVRESP ERROR (… [File: ; Start date: ; Stage: 2]):
check_channel; units mismatch between stages,
skipping to next response now
the problem is that between stages are changing the units in your RES file Ex:
B054F04 Stage sequence number: 2
B054F05 Response in units lookup: V - Volts
B054F06 Response out units lookup: COUNTS - Digital Counts
(…)
B054F04 Stage sequence number: 3
B054F05 Response in units lookup: COUNTS - Digital Counts
B054F06 Response out units lookup: COUNTS - Digital Counts
If the highlighted ones are not the same, that is causing the problem.